Hi everyone, I have some questions now:
The next step now would be to calibrate the energy right? Do you have any suggestions for an easy to come by source for the calibration? Thanks in advance Best regards, |
Beta Was this translation helpful? Give feedback.
Replies: 1 comment 6 replies
-
Hi Dennis! Glad you got it running, here are some answers to your questions:
Yes, the next step would be the energy calibration. Potassium is a good source, however you need a lot of it and measure quite a long time, because the detector sensitivity decreases a lot the higher energy you go. I would advice against any kind of uranium sources, because usually they have a lot of other decay products in them. These peaks will overlap and make it very difficult to assign them to their energy. Maybe you can get your hands on some old vaccuum tubes with Cs-137 or ionizations smoke detectors with Am-241. You can also buy lutetium oxide to measure the natural isotope Lu-176. It's just a metal, but you can also grab yourself some LYSO scintillators (on ebay) that also contain the same isotope. Those come to my mind now. Hope that helps! Cheers. |
Beta Was this translation helpful? Give feedback.
Installing the "clean firmware" did the trick, thanks 😄 The culprit is the LED on the Pi Pico W Board. With the stock firware it worked and this was the only change I did. Commenting out the LED in my version also lead to a correct spectrum. gpio_set_slew_rate doesn't work with the LED on the "W"-Board because the LED is controlled through the wireless chip. Apparently it is driven so hard that this causes some EMI interference.
The spectrum now looks like this:
My guess is that adding white PTFE-Tape to the surface of the crystall not covered by the sensor should move the peak of the distribution further right.
Only thing that looks strange is that the spectrum has a saw-like pattern:
…